Notes on Meeting in Baltimore. Nov 12-13, 2001
Our discussions focused mostly on optimization issues in IMRT:
Intensity-modulated radiation treatment planning. We found many areas
that require further investigation, and in which improved optimization
techniques could have a large impact on practice.
The main treatment device we discussed was a linear accelerator
equipped with a multileaf collimator (MLC). Terminology for various
treament schemes associated with this device is roughly as follows.
Terminology
1) Step-and-shoot.
The beam is aimed at the tumor from a number of
different angles (say, 6). At each angle the MLC is set to a number
of different shapes (say, 3) and a beam of a certain weight is
delivered. The allowable shapes are dictated by the physical
characteristics of the MLC (see below). The shapes used at each
angle may or may not conform to the shape of the actual tumor.
2) "Conventional".
Here, the beam is aimed from a number of different
angles, but the shape of the MLC aperture is chosen to be conformal
to the "beam's-eye view" of the tumor. The beam may be filtered
through a wedge that attenuates the radiation across the beam, in a
fashion that can be modeled by a fixed angle (orientation of the
wedge) and an affine function. Usually, one beam is delivered from
each angle. Sometimes, two beams are delivered (one with and one
without the wedge, or with the wedge oriented in two different
ways). A scheme of this type is the most popular one in current
use.
3) IMAT.
In this scheme, the gantry is moved during delivery of the
beam. The aperture shape may be varied continuously during the
move. It has been found that IMAT can be approximated by a
step-and-shoot procedure in which beams are delivered from angles
along the path separated by about 10 degrees. Hence IMAT can be
modeled and optimized by similar techniques to IMRT, except that
there are additional constraints on the aperture shapes, due to the
fact that the speed at which collimator leaves can move between
successive approximating positions is limited.
More details
An additional degree of freedom is that there may be two to four
different beam types that can be used (electron and photon, different
beam energies.)
Data for the model consist of "dose matrices" associated with each
pencil beam. Each such matrix indicates the total radiation delivered
to each voxel in the target area by a "pencil beam" of unit weight
with a given orientation. The amount of data for a real target is
potentially huge: There are approximately 100 pencil beams that can be
delivered from each direction, around 36 candidate directions, and up
to four beam types. For the 20-cm diameter water cylinder, used for
algorithm prototyping, the dose matrices for a single direction have
been precalculated. A rotational transformation can be applied to
obtain the dose matrices for other angles. For a "real" target, the
Monte Carlo technique must be used to calculated complete dose
matrices from every candidate direction. For some parts of the
modeling and optimization, it may be possible to use approximate dose
matrices obtained via a much cheaper calculation (TERMA ??).
The objective is essentially to deliver a prescribed dose to the
target area, while avoiding overdosing of nearby "sensitive
structures" and of surrounding normal tissue. Treatment planning aims
to achieve this goal by choosing the variables in the model
appropriately. These variables vary according to the type of treatment
(see above) but may include the following:
- angles at which the beam is to be delivered;
- weight of each beam;
- aperture shapes at each angle;
- use of wedge and its orientation;
- energy of beam to be used.
In current practice, the objective function often takes the form of a
weighted least-squares deviation from target dose values. Constraints
are sometimes imposed on dosage in a given region. In addition, there
are constraints of the form "no more than 30% of the target can
receive a dose of more than 20 Gy" that are easy to specify via the
dose-volume histogram (DVH). These are however difficult to model from
an optimization viewpoint.
Total Body Irradiation
Another somewhat different problem we discussed was total-body
irradiation (TBI) used in treatment of leukemia patients prior to bone
marrow transplants. In the perferred mode of delivery, the patient is
laid flat and a wide beam is scanned over him from toe to head. Since
the distance of aperture to the point of incidence of the beam on the
patient's body changes with angle of the beam, and since the volums of
body exposed to the beam varies with body width/thickness at the point
of incidence, the amount of radiation delivered to the body varies
with angle. The objective is to vary the rate of scan so that a
uniform dose is delivered to all parts of the body.
State of Art
The recent paper [1] optimizes a step-and-shoot plan by fixing 6 beam
angles, then choosing 3 aperture shapes and beam weights to be
delivered from each angle, via a simulated annealing
approach. Candidate changes in each variable were rejected immediately
if they resulted in infeasible MLC apertures. The technique for
handling DVH constraints was to change the objective function between
iterations, applying higher weights to those voxels in which dose
"just exceeds" the preferred limit in an attempt to force the right
proportion of them to drop below the preferred limit. The algorithm
resulted in a very effective plan, involving many fewer MLC aperture
shapes than a plan developed by a commercial vendor. Since 7 seconds
are required on the equipment in use at UMB to change MLC leaf
configurations, there is a large reduction in treatment time.
Michael reported on a leaf sequencing algorithm developed by Boland et
al. It overcomes the difficulties associated with specifying allowable
aperture shapes by defining a network that essentially hard-wires the
allowbility constraints. Flows through the network define valid
configurations of the MLC leaves. A prescribed dose can be solved as
a network flow problem. Michael has a simple implementation in
GAMS. The objective in this problem is the total dose delivered, NOT
the number of different shapes used. A depth-first search can be used
to "back out" the different shapes from the solution of the network
flow problem.
 [1] D. M. Shepard, M. A. Earl, X. A. Li, S. Navqi, C. Yu, "Direct
Aperture Optimization: A Turnkey Solution for Step-and-Shoot IMRT,"
2001. |